Front Page › Forums › MMSIT › MMIST – Problem images are not complete › Re: MMIST – Problem images are not complete
December 6, 2005 at 12:05 pm
#958
Markus Stoeckli
Moderator
I think there is a fairly simple solution to this problem. In the ‘Instrument Settings’ of the ‘Control Panel’ you must disable the ‘Automatic Voltage Shutdown’. This is typically set to 30 minutes and then the voltage is switched off – for this reason the top half of the images remains black.
markus