I am using an Bruker Ultraflex II, and am having problems looking at my data in BioMap.
After running anlyze this on the data, many mass spectra are distorted after importing them into Biomap. They seem to max out at (32k), and the intensity goes between 0 and max. I have tried to not bin the data and the same problem is still encountered.
Any suggestions?
Post edited by: Heien, at: 2006/11/02 22:53
Post edited by: Heien, at: 2006/11/02 22:54
Yep that sounds familiar :), it is a bug in the peak intensity conversion in Analyze This – apparently it only shows up when working with smoothed or background subtracted spectra.
I’m currently finalising an updated version of Analyze This in which the bug is fixed and hope to post this to this website in the following week; in the meantime you might try converting non-manipulated spectra and see if that helps.
cheers,
stefan
Thanks for the info – I will look for the update.
Micahel